Strain, stress

NOTE: In Strain & stress it is important to correctly setup the Analog IN.

Introduction

In this section we will cover how to setup the bridge operations like:

  • Balance
  • Reset
  • Short
  • Shunt
  • Shunt cal check
  • Compensate

It also explains bridge sensors with TEDS interface.

Set up the sensor (with TEDS)

For using the benefits of automatic sensor detection (automatic reading of scaling, serial number, …) please activate the check-box “DSI adapters/TEDS sensors” in DEWESoft -> Settings -> Hardware Setup:

Strain\&Stress_Setup_TEDS settings

Start DEWESoft; here the sensor with TEDS (a tuning fork with strain gage applied + TEDS chip inside sensor connector) is plugged into channel 5:

Strain\&Stress_AnalogIn

All the settings like “Measurement Type”, “Bridge mode”, “Excitation” and “Scaling” are read out and set automatically. Note the grayed, locked fields, which cannot be changed (at first):

Strain\&Stress_AnalogIn_Channel setup

Balance the bridge

From the screenshot above you see that the current sensor offset (stored in the TEDS) is 0 %. In the scope screen above the scaled value currently read (right side) is 2027 um/m. That is because the strain gage does not have exactly 350 ohms, which is normal, due to mechanical tolerances.

Click on “Balance” to zero the bridge.

Strain\&Stress_AnalogIn_Channel setup_Sensor unbalance

The unbalance will be measured and shown. If it is out of limit, the field will change background color. This is the indication to check the wiring or a strain gage itself.

Strain\&Stress_AnalogIn_Channel setup_Sensor Balance

The input signal is now 0 mV/V.

Strain\&Stress_AnalogIn_Channel setup_Sensor Balance_Scope

Balance is done. That’s all! Now you can START your MEASUREMENT! To perform it on a number of channels at the same time, use the >> Group operations.

Sensor unbalance limits SIRIUS-CD, SIRIUS-HD

If unbalance value has yellow backround it means that the selected unbalance value is >500% of the Selected range. On the pictures below the voltage range is 2mV/V so in this example it becomes > 10mV/V e.g. more >500%.

If unbalance value has red backround it means that the selected unbalance value is >100% of the HW range. On the pictures below the voltage range is 10000mV/V and the HW range is 5000mV/V which results in 100% of the selected range. Note that full range includes offset and selected range.

Sensor unbalance limits SIRIUS-HS

The specification for HS-STG in the SIRIUS manual says: Sensor Balance Range …. 220 % of Range (70 % for input Range ≥5 Volt)

This means it will turn red if the unbalance is >= 220% of Range, where in this case the real Range is taken. This is true for all voltage ranges below 5V, so 5V and above the unbalance it specified with 70%.

The spec. says 220% of Range is allowed by HW, so 10,76311mV/V * 2,2 (220%) = 23,678292 mV/V max allowed unbalance. And now also the unbalance offset can do a bit more than the 220%.

Reset the balancing

Click on “Reset”. The sensor offset is removed, initial state…

Strain\&Stress_AnalogIn_Channel setup_Sensor unbalance_RESET

… and you see the unbalance of the sensor again.

Strain\&Stress_AnalogIn_Channel setup_Sensor unbalance_Scope

“Balance” and “Reset” are opposite operations.

Short

The Short and Shunt operations are mainly thought for checking if the connected strain gage and the measurement amplifier are OK. When using the Short operation, the pins 2 and 7 (the input pins of the amplifier) are internally shorted and you see the offset of the amplifier.

First perform a “Reset” to cancel the stored offset.

Strain\&Stress_AnalogIn_Channel setup_Sensor unbalance_RESET

Then click “Short on”.

Strain\&Stress_AnalogIn_Channel setup_Amplifier_Short on

If the output (see left side) is close to 0mV/V, the amplifier is OK.

Strain\&Stress_AnalogIn_Channel setup_Scope_Shunt ON

Disable the short again by clicking “Short off”.

Strain\&Stress_AnalogIn_Channel setup_Amplifier_Short OFF

Shunt

The Shunt operation is thought for checking if the connected strain gage is OK. From the wiring schematic in DEWESoft X you see, that the amplifier already comes with the integrated shunt resistor.

Strain\&Stress_AnalogIn_Channel setup_Shunt scheme

The idea behind is to “shunt” a resistor of known value parallel to one resistor of the bridge to achieve a known, calculable unbalance. With the “Shunt calibration” we can automatically check the measured value against a predefined one (from sensor database or TEDS).

For the measurement, this internal resistor is disconnected again, of course.

Let’s check out the formulas:

Strain\&Stress_AnalogIn_Channel setup_Shunt scheme MODEL

(“Shunt calibration for Dummies; a reference guide” by LaVar Clegg, Interface Inc.) (The constant 250 represents $\frac{E}{4}$, where the excitation voltage (E) is assumed to be 1000mV (1V). This standard assumption is used because the bridge output is generally expressed in mV/V.)

R3 and R4 are part of the bridge completion, internally of the amplifier. The upper Rb is the connected strain gage, the lower Rb also internally of the amplifier together with the shunt Rs.

With a 350 ohms quarter bridge and an internal shunt resistor of 59,88 kOhm SIRIUS STG module) the expected unbalance should be: Vs = $\frac{250}{59880/350 + 0,5}$= 1,45699775 mV/V

First click on “Balance”, because the formula is only valid on a balanced bridge (both Rb = 350 ohms).

Strain\&Stress_AnalogIn_Channel setup_Sensor unbalance

Then click “Shunt on”.

Strain\&Stress_AnalogIn_Channel setup_Amplifier_2 Short on

The output value comes very close to the expected value (1,4566 mV/V).

But how big exactly is the error?

Shunt cal check – preparation

Shunt cal check is done on scaled values, therefore we take a look on the strain scaling. The max input from the graph above is 10 mV/V; the max scaled output signal is 20 000 um/m. So, the scaling factor is 2000. Our target value of 1,45699775 mV/V would equal (x 2000) = 2913,9955 um/m.

In the right upper section of the channel setup window, unlock the TEDS settings (or settings of the sensor database if you don’t have TEDS).

Enter the calculated 2913,9955 um/m in the field “Shunt cal target” and press “Enter”. The resistor is OK with 59,88 kOhm.

Then switch to the “Edit sensor” tab and click “Write to TEDS” (or “Save Sensor” if you don’t have TEDS). Of course this procedure has only to be done the first time.

Strain\&Stress_AnalogIn_Channel setup_Sensor_Write to TEDS

Shunt cal check

Now perform the “Shunt cal check”:

The result is shown, and looks very promising in this case (-0,1 %). Strain gage is OK.

If the value is bigger than 2%, it will become red.

Adjusting the 2% limit is now possible in DEWESoft: go to Settings -> Advanced -> Hardware -> Amplifiers -> Shunt calibration error limit [%].

Strain\&Stress_AnalogIn_Channel setup_Sensor_Shunt CAL ERROR

Shunt cal check – with custom shunt

What do you do if the expected shunt cal result is already calculated in the datasheet / calibration sheet of the sensor, but for a different shunt resistance, e.g. 100 k?

You can just use it!

For this example, we take the well known formula from the “Shunt” chapter, and calculate the shunt result with 100k shunt resistor. Vs = $\frac{250}{100000/350 + 0,5}$ = 0,873 mV/V, with scaling of 2000 -> 1746,943 um/m.

Enter the values as shown below. When entering the resistor value, the field gets yellow. Please confirm the value with pressing “Enter”.

Strain\&Stress_AnalogIn_Channel setup_Sensor_Custom shunt

Then perform the “Check”.

As you see, the software automatically takes care about different resistor values. The asterisk (*) symbol after the result indicates that the check has been performed with a different resistor value than physically available.

NOTE: Hovering over the Shunt cal target bracket will display the shunt cal target values based on the possible shunts of the amplifier.

Strain\&Stress_AnalogIn_Channel setup_Sensor_Shunt CUSTOM SHUNT

Lead wire compensation

If you have a long cable to the sensor and “quarter bridge 3 wire mode”, DEWESoft can also cancel out the wire resistance. (In the other modes the wire loss is canceled already because Sense and Excitation lines are connected directly at the sensor).

Here you see a quarter bridge strain gage with 3-wire connection. In each of the 3 lines we have built in a resistor of 11 ohms for showing the principle. Click the “Compensate” button.

The correction factor is calculated (1,031 in our case) and the measured resistance (10,8 ohms) displayed. (Please take into account, that this function is only available, if the Shunt resistor inside the module is connected to the Exc+ line, which is the default).

Zero

There is also a function called “Zero”, let’s look at the difference to “Balance”.

Strain\&Stress_Zero_Balance

Imagine, we have a force transducer with strain full bridge output. It will measure the weight in our experiment. In the first picture, we measure just the unbalance of the bridge sensor, e.g. 35 N. Let’s do a “Balance”. The output is now 0 N. A vehicle is put on the test bed. We measure the weight, it is 12 000 N. For our measurement only the changing of the weight is of interest (e.g. vibrations of the vehicle), so we cancel out the fix offset with the Zero function.

Strain\&Stress_AnalogIn_Zero

Click the “Zero” button in Channel Setup (can be reset by right mouse-click). The output is now zero again. Note, that this is a pure software subtraction. If the range was set to “Automatic”, the range is automatically adapted to -52000…+28000.

The range can be set to “Automatic” in the channel setup window of the appropriate channel (right mouse-click).

Strain\&Stress_AnalogIn_Channel setup_Set to automatic

Now all offsets are canceled and we start the measurement. After the measurement you can go back to channel setup and remove the Zero. If it still shows the same weight, you can be sure, the sensor is still OK.

This function can also be accessed in the Measure mode (but NOT while storing!), see >> Group operations.

Group operations

On top of the Channel setup there is a bar with channel actions named Channel actions.

Strain\&Stress_AnalogIn_Channel actions

Depending on which amplifiers are used, more or less buttons are visible. Clicking on a button performs an action on ALL the channels.

For the following steps, please activate some more columns on the channel setup table. Click on one of the small icons, and select “Edit columns”

For information about Channel grid see -> Setup -> Channel grid.

Select the fields “Sensor check target”, “Sensor check result”, “Sensor check error” and “Group”

Strain\&Stress_AnalogIn_Choose columns

Group – Bridge balance

In the custom SIRIUS slice below there are 2 STG modules, each with a sensor connected. Click on the “Balance sensors” button.

As you see, all bridges are balanced with one click, input values become 0.

Group – Shunt Cal Check

From the picture above, note that there has to be a ShCal Target entered before doing the check (1,45699775 * 210 * 2 = 611,939 N/mm²; see section “Set up the sensor”).

Now click on the “Shunt cal check” button.

After a few seconds, you see the result in the additional columns.

Group – Balance amplifiers

The “Balance amplifiers” operation is applied on all channels at once. It is used to correct the amplifier offset before measurement, e.g. the drift on long-term measurements, or if the device is put to extremely different temperature conditions in short time (e.g. performing a measurement at -20°C, and immediately after that at +40°C). Depending on the usage, it does not have to be done before each test, generally only from time to time.

  1. Let’s assume we have a SIRIUS STGM amplifier in bridge mode, set to the smallest range of 2mV/V. When there is nothing connected, we will see only noise, so please click the “Short on” button. The IN- and IN+ will be shorted internally to determine the amplifier offset, in this case 51,1 μV.

  2. Then exit the channel setup and do the “Balance amplifiers”, it takes some seconds.

  3. Repeat step 1 and check the amplifier offset again. In our example it is only 0,05 μV.

  4. If you want to reset the offsets again, please go to Settings -> Hardware Setup -> Analog -> and click the “Set card…” button on the right side. In the tab “Ampl. Balance” you see a list of all the applied offset corrections for each channel. 

Group operations with different groups

In the Group column assign one channel to “Group 1”, the other to “Group 2”.

Now you can apply (bridge) operations group-wise.

Group operations in Measure mode

Now, with one channel assigned to Group1, another to Group2, let’s go to the Measure mode.

There are two new buttons called “Zero” and “Bridge”. 

If the “Zero” button is not there, most likely sensors from the sensor database / TEDS sensors are used, and changing the offset is not allowed.

Check sensor settings in Channel setup or Sensor editor. If the “Bridge” button is not there, no bridge amplifiers are set to “Used” in channel setup. 

Now you can do a selective “Balance” on the group you want or on “All chs”. 

Note, that DEWESoft is in Acquisition mode, but NOT storing! When pressing “Store”, of course no balancing is possible during measurement any more.

Shunt/Short at beginning and end of measurement

At the end of a measurement you may want to check if the strain gage and the amplifier are still OK.

Maybe you also want to see, if the bridge was drifting over time due to temperature or other effects.

Start the measurement, click the Store button. The “Zero” button will disappear, because zeroing also changes the channel min/max limits, and that is not allowed during measurement.

Also balancing the bridge is not possible at this state any more. Do a “Short on for 1s”, wait a little bit, then press “Shunt on for 1s”.

At the end of the measurement – when you are still storing (!) – do again a “Short on for 1s” followed by a “Shunt on for 1s”.

Stop the measurement and go to Analyse mode.

Activate the cursors in the properties of the recorder instrument (on the left side).

Move the white Cursor I to the Short position on start, and Cursor II to the Short position at the end (gray arrows). You can also lock the cursors to not lose them when zooming in and out of a longer measurement.

On the right side the Delta will be shown. In our case it is 0,0 – measurement OK.

Then move the cursors to the Shunt positions (black arrows), repeat the procedure to also check the Delta of the Shunt measurement.

HINT 1: Strain gage configurations in DEWESoft X

Until now it was rather difficult to select the proper strain configuration. You had to find out the bridge factor in the tutorials table and look for materials coefficients.

Strain\&Stress_Hint_Layouts

Depending on the selected “Bridge mode” in the amplifier settings (Full, Half, QuarterBridge), it will now graphically show a various number of possible bridge configurations, and furthermore, you can select the material for using the correct Young’s modulus.

Another error source is eliminated.

Strain\&Stress_Hint_1

Strain\&Stress_Hint_2

Strain\&Stress_Hint_3

HINT 2: Configuration for strain gage load cells

The load cell used in this example is full bridge. The correct ”Input type” and ”Excitation” voltages have to be set correctly. It is recommended to perform a ”Bridge zero”, be careful that it is performed with no attached load.

The next step is to enter the sensor scale factor. The load cell is calibrated in kilograms and in the specification sheet is stated that it outputs 1 mV/V per 10 kg. The value is entered in the scale factor and this concludes the setup.